I've been working with a shared Waters 2695 with a 2996 DAD for the past year, and another user pumped 0.2M phosphate buffered mobile phase through the instrument (with their column) about 6 months ago. This change in mobile phase originally manifested as over-pressure issues, but copious flushing with warm water and isopropanol—and backflushing my column—eventually solved those issues & the instrument has run without fault for the past 4 months. However, I am now getting a "Plunger Homing Fault (0)" error popping up twice during the instrument's self-initialization. Once before carousel initialization (which can be dismissed) & permanently afterwards. Unfortunately the initialization is required for any direct control of the instrument. This particular error message is not described in the 2695 Separations Module manual, however the Waters support website describes this error appearing after using buffered mobile phases. Despite the Waters support article, I'm not inclined to believe residual buffers suddenly precipitated out of thoroughly flushed fluidics, 6 months after the buffer was run through the system, and after 4 months of flawless operation without any of the previously observed pressure issues. My first fixes for this error were loosening the actuator vent during initialization, and replacing the inline filter element. Neither of these steps solved the error popup. Other culprits discussed on other chromatography forums seem to point to a faulty F1 fuse on the solvent management system board. I tested the existing one with a multimeter and got very inconsistent resistance readings, but replacing the fuse did not solve the error during initialization. I checked my seal wash line's prime and it appears to not have any bubbles trapped in it. Presumably it's been washing the back sides of the actuators' plungers, however being locked out from direct control, I can no longer re-prime the seal wash line. Some other culprit parts that I've found discussed online are:
Has anyone encountered this error before? What solution worked for you? I'm very much hoping I won't need to bring in a technician to rebuild the primary actuator but thinking it might come to this.
But if in a hurry a workaround is to open the dry purge valve until the initialization is complete, after that wash your system through
I'm glad you think it's the check valve since that is a much easier replacement versus disconnecting & pulling out the torpedo tubes. Did you attempt to clean the check valve by sonicating in water or methanol? Or did you simply replace the part with a new OEM check valve?
I looked through the separations module operator's guide for what the commands to open V1 to V4 are, but for the life of me I can't find it! Pressing the "Menu/Status" button shows the open/closed status of these valves, but I'm unable to use the arrow keys to select any of these icons, unlike directly changing gradient, flow rate, etc. The "direct function" tab accessible from the "Menu/Status" page never mentions valve control either. Do you know what button presses must be navigated to be able to manually open the valves? Thanks for the help!
|
||